loss of compound during HPLC purification
Posted: Thu Jun 21, 2007 6:17 pm
by slentz
I am purifying small molecules by preparative chromatography. My purified yield is usually around 50%.
My question is where does the compound go? If I start with 1g of crude compound, and I end up with 0.45-0.5g, where is the rest going? I know some yield is impurities that get purified out but if my crude purity is 90% that's only 10% loss. I'm sure I get nickel and dimed out of a few more % due to tube transfering etc.
Does some of the compound get lost on the column? How can I increase my yield?
Posted: Thu Jun 21, 2007 7:14 pm
by chromatographer1
Slentz
without knowing the chemistry of the product I doubt anyone can give you an answer that will mean anything.
If your starting material is 90% then by processing it almost half of it is converted to something else. The way in which that happens depends upon the chemistries you are dealing with.
Of course something of your starting material might be left on your column but I would doubt it would be almost 50% of your starting material.
Most likely the material is unstable under the conditions you are performing the purification step and the loss is due to that factor.
Increase the stability of the purification conditions and your yields will increase.
How?,
share the details of the chemistry involve and someone may be able to help you.
best wishes,
Posted: Fri Jun 22, 2007 6:25 am
by Jackus
Hello Slentz
there may be several reasons for material lost.
1. How was the purity of your material determined?
If is determined as area % there may be difference in response factor of product and impurities. It means that content of product was overestimated.
2. Do you collect only peak of product during preparative chromatography or you collect fractions of the same volume and put it together after analysis?
There may be some hidden, undetected material in fractions on the start and end of peak. Did you analyzed all fractions?
3. There may be material decompostition during chromatography but from my experience it is not possible to obtain pure product in this case. Always is contaminated with degradation product.
Flush column with strong solvent at the end of separation and analyse obtained material for presence of product.
Regards
Posted: Sat Jun 23, 2007 8:45 pm
by Bruce Hamilton
I find that I typically lose about 5-15 % when running prep HPLC on fairly stable molecules, but can easily lose more without trying too hard :-) .
It's always worth remembering that preparative chromatography usually means longer time in solvent, and longer evaporation times. The sample quantities will make losses more visible than on analytical columns. If you always loose 10% on an analytical system, you may never notice.
Typically, the two common causes of my Prep HPLC losses are:-
1. Decomposition.
If you know the major decomposition peak, and you see it smearing toward the parent peak, you could have on-column decomposition, which I fortunately don't encounter often.
The more common form is decomposition in the loading solvent or mobile phase, which can be detected by performing replicate injections with a significant time period between and checking peak area. I establish a baseline point by ensuring the first injection is < 5 minutes after dissolving the sample.
The most common source of decomposition losses is concentration of the eluate after elution from the preparative column. I always try to analyse the eluate as soon as it leaves the column, and calculate how much analyte is present. It's often during evaporation where recovery drops from 90% to 50%.
Many molecules croak if they are rovaped down from mobile phases that have modifiers such as TFA present. Ensure temperature and/or light aren't degrading your product. You should check stability to allow for the longer time/temperature enviroment of preparative HPLC solvent volumes.
2. Precipitation
The major analyte isn't soluble in your mobile phase once it's separated from the muck or injection solvent. In preparative HPLC, this is usually accompanied by backpressure increase, and completely-blocked prep columns aren't uncommon.
Having $0.5 million worth of cGMP pharmaceutical product on a column which has almost blocked after 15 hours of increasing back pressure and decreasing flow is fun only if you're a disinterested spectator.
A more polar solvent at the end of a run may not redissolve precipitates. You may have to raise the temperature, make major modifications to the mobile phase ( such as changing to pure ethyl acetate ), and greatly decrease the solvent flow, to achieve dissolution.
Other causes, such as prep columns and solvents being different to analytical conidtions are less common. You should always try to maximise the quality of the preparative column and solvents, as trace impurities can wreck havoc when large volumes are concentrated.
Always ensure all your samples are actually loaded onto the column, and not being left behind in flasks or injector loops, by binding. I've also managed to fill loops in the injected position, and watched the sample run into the injector waste line. If you have large losses, analyse injector wastes in case a valve ( or operator :-) ) isn't working correctly.
Please keep having fun,
Bruce Hamilton
thanks
Posted: Mon Jun 25, 2007 4:10 pm
by slentz
You've all given me plenty to think about. I appreciate all your input.
Thanks!
Posted: Tue Jun 26, 2007 8:23 am
by Ary
Without wanting to teach anyone how to suck eggs I thought the following maybe useful.
A quick method we use to check recovery is to take 100mg of Imipramine and dissolve it in 5ml of HPLC grade water. From this we take 900ul aliquots and inject them onto the prep system. The collected fractions are collected and pooled if necessary and made up to 100ml with water. 20ul of these solutions are then injected onto an analytical column. As a standard we take a 900ul aliquot from the original solution and make this up to 100ml and assign it an arbitary purity of 100%. By comparing the area response of the collected fractions with this standard we then calculate our recoveries.
MOST of the time our recoveries remain reasonalbly constant even as chemistries change, unless we run into some of the problems already mentioned.
by Ary