only get one good injection
Posted: Fri Jun 15, 2007 9:15 pm
I have a separation that is just a mess. I am trying to do a separation of opiates. I have Waters 2790 and Quattro micro. I am trying to use an Agilent SB phenyl, 1.8 um, 50 x 3.0 mm. I am doing a gradient separation. No guard column for this column, they recommend using a 0.2 um frit in a holder.
I am doing a gradient separation. I make one injection, things look OK. I make a second injection, my early eluting peaks spread out and start to split, but later eluting peaks look OK. This gets worse and worse until I change the frit, and/or flush the column, then I get one more nice injection. Also, early eluting peaks are about twice as wide as later eluting peaks. This seems odd to me.
I have made fresh mobile phase and flushed everything out. Mobile phase is ACN:water:100 mM ammonium formate pH 3 (yes I know it would be better to mix the buffer in with the water and ACN and not have the pump do it, but I inherited this set-up). We have about a dozen other isocratic assays that use this configuration and they are fine.
I hold at 5:93:2 ACN:H2O:buffer for two minutes, ramp to 25% ACN over two min, hold 0.5 min, then ramp to 95% ACN and hold for 1 min, then back to beginning, with 3.5 min re-equil time. I have a 600 uL pre-column volume, which is the dumb name Waters uses for gradient delay (sorry Waters people, it took me two days to figure that out).
Any ideas what is going wrong?