We have encountered a problem in which the detector fails our wavelength calibration (phenanthrene & anthracene in the cell), but when the internal wavelength check on the instrument is done and there are no error messages. I guess that my question is, what is the value of the self check?
George Balock

I can't remember - are 2487s dual lamp detectors? If so, the UV lamp may have gone bad. As HOlmium oxide filters are quite useful in checking wavelengths in the visible end of the spectrum, they are not so good at checking the UV end... hence the passing of internal checks.
Thanks,
DR
Image

DR – one lamp only.

Gbalock – I guess it’s a question of acceptance criteria. Try to compare the tolerance allowed in the factory settings to your procedure’s limits.
The need for wavelength accuracy can vary a bit, depending on the intended application, but as a rule of thumb 2-3 nm should be acceptable.

Best Regards
Learn Innovate and Share

Dancho Dikov
3 posts Page 1 of 1

Who is online

In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry