Problem with C18 column on preservatives
Posted: Tue Jun 12, 2007 1:53 pm
Hello everyone!
I work for a very very small company wich distributes VARIAN consummable.
Although I'm not a specialist in chromatography(studied a long time ago ), I try to help one of our customers which has a problem with his chromatogram.
He works in a cosmetic company and needs to establish a procedure for the quantification of preservative in his products.
For now, he's working on conservative (especially parabens) as a raw materiel, so its pure.
He had a problem with his former column (a VARIAN Microsorb 5µm C18 4.6x150 mm), which disapeared after using a new column Microsorb 3µm C18 4.6x100mm) : the peaks became wider, and he didn't manage to separate n- and iso- butyl paraben
Recently, the same problem appeared again, with both columns (he reinstalled his first column and chromatograms were OK). He sent me a chromatogram from his results:
He injected the same sample for 3 days, without changing the mobile phase, and without using any buffer.
Each color reperesents the chromatogramm of the day (during 3 days :day1 black, day2 red, day3 blue)
Focus on the n- and iso- butyl paraben
The method he uses is :
Column Microsorb MW 100-3 C18 100x4.6mm, oven temperature=40°C, Flow= 1 ml/min
0 to 6' : H2O/MeOH 60 :40
6 to 10':H2O/MeOH 60 :40 to 40 :60
10 to 18' :H2O/MeOH 40 :60
18 to 19' :H2O/MeOH 40 :60 to 60 :40
Is it normal that the 3 chromatograms showed different retention times, and do you have any idea why the resolution between n and iso butyl paraben decreased?
Thank you very much in advance for you help!
Bye
PS: sorry but my english is not perfect
Bye