Zero pressure on 1100 pump.

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
We're running an 1100 with a quat pump for about 18 months with no issues. However today the pressure suddenly died on us.

Yesterday we were running normal phase, (Ethyl Acetate/Hexane) flushed out the system with IPA before running reverse phase with changed to a 50:50 water:methanol system which was fine for a few hours. We then changed to a 20:80 phosphate buffered (pH7) water:MeOH system (first time we've tried it) and within minutes we'd lost all pressure, no matter what flow rate we try.

I'm guessing that the buffer is responsible. Which begs the questions why and how do we fix it?

Matt

Your seals became swollen in the NP mobile phase and conformed nicely to the pistons, but then shrank in the RP mobile phase. There are different seals for RP and NP; get the right kind. You can get away with using the wrong kind for a quick project, but not for extended use.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 14 users online :: 2 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry