USP Residual solvents chapter <467>

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Has anyone been able to successfully analyse USP <467> on an Agilent 6890 GC/headspace, using the method as it is written in the current issue (USP 30)? I am using a Restek 624, 0.32mm, 1.8um, 30m, Helium, split ratio 5:1. I am finding it hard to identify the Class 2 Std A and Std B peaks (without doing a heap of upfront work, or injecting the solvents individually!) I have tried the test for water soluble articles, procedure A, using the operating parameters suggested in the table (conditions 1 and 2).
Many thanks!
:roll:

Without using an MS, I don't know how you might identify the compounds without injecting individually.

I am runnig these by GC/MS and am only having issues with pyridine and nitromethane.

I'd give you the elution order on my column, but I fear that my employer may flay me :P

Nice application note, thanks. Notice the peak shape of the pyridine and the sensitivity of the 1,2-dimethoxyethane :roll: :lol:
4 posts Page 1 of 1

Who is online

In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry