New problems with indole skatole

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hooray for the forum!

I'm long pass my problems with my exceptionally high baseline, thanks for everyone who helped. But now new problems have arisen.

I know it's a contamination but where? I've cleaned or replaced soon the entire machine. The problem is, that when I inject only the solvent (chloroform) I get peaks, as if I'd injected a poor sample. With a blank injection, no peaks appear.

I've replaced the needle and the inlet, cleaned the gold plate and the detector. My column is almost brand new.

Where to look next?

Specs for my setup:
- Capillary column (Restek Stabilwax DA)
- Solvent is chloroform, needle cleaning solutions chloroform and MeOH

Heikki,

It sounds like your chloroform is contaminated, or that it is leaching out something from your vial septum or your injector inlet septum.

best wishes,

Rod

I suspected as much, and changed my chloroform and the inlet septum. It could be the vial septums though...

Try hexane or dichloromethane.

Best regards.
4 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry