akthmps » Mon Jun 11, 2007 5:26 pm
Here is what I now observe (at 80C oven, 1.0mL/min flow, 250C injector). I changed out the septum and did some basic injector maintenance. I performed an air/water check. Air looks fine, water level is massive. The injector is in splitless mode. I then turned on the injector splitter to 100:1, within 30 seconds the water level fell to "normal". I turned the splitter back off, and within a minute or so, the water level was once again very high. So, the water contamination seems to be a function of the injector split state. Any suggestions?
I did a septum purge calibration this morning. At 10 psi, the Varian manual states the purge flow should be around 4.5mL/min for the 1079 PTV injector. The flow was very close to 4.5mL/min, I did a minimal adjustment and saved the changes.
Why should the split state be affecting the water levels in my system? Any suggestions? I really need to be doing splitless injections...
Thanks,
Aaron