Phenol Disapearance
Posted: Wed May 30, 2007 5:01 pm
Hi all. I am analyzing phenol in water from 100 to 2000 ppm using A DB-5 capillary column, split injection to an FID. My instrument is a HP 5890. with auto-injector.
After external calibration and a couple of sample runs, my phenol peak has disappeared (std and sample). I have changed the glass liner, septum, cut the column, baked out the gc, checked all flows and temperatures/settings, reduced the split. I have run other test analytes such as acetone and ethanol and the chromatography is great.
I am now having the student mix a 1 % test sample.
My temperatures are: injector 175, column 90 isothermal for 6 minutes, detector 250, injection volume 0.5 ul. The phenol was tailing quite a bit as well.
I am lost on this one. Help please!
Ricardo
Ricardo