detection limit with FID

Discussions about GC and other "gas phase" separation techniques.

5 posts Page 1 of 1
I run some standards with very low concentrations of propylene oxide and I got a linear correlation of the area vs amount of p.o. (with FID-range2) According to the equation of the line the detection limit is 1,5*10^-7 g. Is this normal? or is very large number?

Tolisfoto,

Why range 2? Are you using Chemstation or some other data system and the analog out on the 5890? Also, what level of standards are you using and then are you extrapolating? FID is typically low nanogram on column detection limits for many organics.

AICMM, I don;t have any datasystem.. I used standards in the range 0.8 - 8 μg P.O and range2 because I use this range for the samples..

Are you by any chance doing split injections ? If so, what is the split ratio ?

Peter
Peter Apps

peter I don't have split injection.. I use packed columns.. However, considering that in range2 the sensitivity of the FID is 100 times smaller than range0, I think this detection limit is normal
5 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry