How do you set up an injection delay on a Waters Alliance?
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
6 posts
Page 1 of 1
Have a Waters Alliance and Quattro micro and would like to set up an injection delay. Can't find anywhere to do it.
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Hi sjmarin,
I assume you’re using Waters CDS e.g. Millennium or Empower. If yes, then just enter the desired time (in minutes) in the “injection delayâ€
I assume you’re using Waters CDS e.g. Millennium or Empower. If yes, then just enter the desired time (in minutes) in the “injection delayâ€
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- Posts: 580
- Joined: Fri Mar 16, 2007 11:29 pm
I'm using Empower not Masslynx with the Alliance.
With Empower you have a designtated field in the "Instrument Method" where you can enter it.
@danko: I'm a bit curious about you suggestion.
Do you mean the "Next Injection delay"? This is just a "post-run" delay time without data acquisition.
As I'm writing this, I realize that sjmarin's question is not fully clear.
Sjmarin, which delay would you have? A post run delay without data acquisition (e.g. reconditioning the column) or a "real" inj delay for the correction of gradient delay time?
With Empower you have a designtated field in the "Instrument Method" where you can enter it.
@danko: I'm a bit curious about you suggestion.
Do you mean the "Next Injection delay"? This is just a "post-run" delay time without data acquisition.
As I'm writing this, I realize that sjmarin's question is not fully clear.
Sjmarin, which delay would you have? A post run delay without data acquisition (e.g. reconditioning the column) or a "real" inj delay for the correction of gradient delay time?
-
- Posts: 1408
- Joined: Thu Jul 28, 2005 2:08 pm
Hi Hollow,
Yes I meant “Next Injection Delayâ€
Yes I meant “Next Injection Delayâ€
Learn Innovate and Share
Dancho Dikov
Dancho Dikov
-
- sjmarin
I would like to start my gradient, and delay injection of my sample until some time after the gradient has started.
I have also found something in MassLynx that is called "Pre-column volume". It says "enter the volume of solvent to pump through the column between the time the gradient starts and the time of injection". That sounds like what I want to do, so I'll try it.
Thanks. This software does not seem very intuitive to me and it doesn't seem easy to figure out how to do things that should be relatively simple.
I have also found something in MassLynx that is called "Pre-column volume". It says "enter the volume of solvent to pump through the column between the time the gradient starts and the time of injection". That sounds like what I want to do, so I'll try it.
Thanks. This software does not seem very intuitive to me and it doesn't seem easy to figure out how to do things that should be relatively simple.
6 posts
Page 1 of 1
Who is online
In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science