danko » Mon May 21, 2007 7:40 pm
Hi semivoadude
It would be helpful to know if the peak/s in the blank had the same retention time as the main peak in the standards/samples or are they just appearing randomly/non systematically in each blank injection?
Here are my thoughts about the described setup: Water alone is not the best solvent for proteins. Usually some salt (in very low concentration e.g. 0.01 M) is needed in order to salt-in the protein. So your needle rinse is not the best choice. The salt solution mentioned above wouldn’t be my first choice for needle rinse either. Here is a suggestion for needle rinse I would put to a test: 0.04% phosphoric acid in water plus 10 – 20 % isopropanol. As you can guess I am suggesting to you that the problem might be carry-over due to inadequate injector wash.
You don’t mention the phosphate concentration in solvent B, but if it is high e.g. 1 M or something like this, there is the possibility of salting your protein out, due to the abrupt change of solvent from 0 to 100 % B in a single step. At least some of the protein. The salted-out protein can subsequently be dissolved and elute in the following run as carry-over.
The difference in pH between the solvents can often be a very difficult parameter to control and is unadvisable. It can also cause solubility problems.
Finally to your question:
I heard flushing at a low flow rate actually can "clean" the system more efficiently, is this true?
To the best of my knowledge it’s not true. And I can’t think of a rationale for this assumption.
I’m not saying that flow-rate of 2 mL/min is the best choice, and it’s probably not – even more so in ion-exchange mode. But the column diameter should be taken into consideration in order to suggest a different flow rate. So I’ll leave it here.
Best Regards