Dear All,

I would like to know if it is right to add the peaks at 1, 2 minutes (in HPLC) in total impurities, if no such peak is there in blank or placebo.

Regards
jUST dO iT....

I would say yes as impurities can be polar compounds that can be eluted very early in chromatography (assuming reverse phase chromatographic conditions). The fact that you do not see these compounds in blank or placebo reinforces the likelihood of these being impurities of your compound of interest...

I would very much agree with Kostas and I’d like to add some thoughts too.
Peaks eluting at 1, 2 min hence most probably at void volume, are not that easy to quantitate. They tend to be very narrow which demands higher sampling rate of the detector. It doesn’t need to be a problem now-a-day with modern detectors, but it is rarely considered by the users, so narrow peaks are seldom defined correctly (15 – 20 sampling points per peak)
Another problem I have experienced with early eluting peaks is that they tend to be very high and in combination with their sharpness it is very difficult to see if they are off scale.
I guess the right approach here is to optimize the method in order to retain and potentially separate early eluting peaks which will facilitate their quantitation greatly. Maybe in this situation it’s worth trying stationary phase with embedded polar groups.

Best Regards
Learn Innovate and Share

Dancho Dikov

Is UV detection assumed here? With all the possibilities of refractivie index, scattering, etc., effects up front, it would be advantageous to have a wizzard to do the quantitation?

I'll elaborate on HW Mueller's comment: attempting quantitation on any peak eluting at or near t0 is unwise. Your system is in a state of "dis-equilibrium". In general, you should have a k' of at least 1 (i.e., retention time of at least twice t0) for good quantitation.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

One should be careful about the "placebo " supplied by Formulators, Some of them, the placebo is just a mix of ingredients and they do not bother to process that mix through the same process as of formulation with drugs. Some specific points are,

1. No additions of buffer, pH modifiers in placebo.
2. No compression cycle ( including solvents used) followed for placebo tablets.
3.No use of gases
4. Reaction between placebo components with drug.

While doing method development for impurities , we always ask for placebo produced with same process except addition of API.

JM
6 posts Page 1 of 1

Who is online

In total there are 34 users online :: 2 registered, 0 hidden and 32 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 32 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry