[M-2] fragments in peptide LC-MS-MS

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
Hi,

I'm trying to run protein digests on my Bruker HCT ion trap, with auto-MS2. The chromatograms look ok (as far as I can tell!), but I'm getting really poor quality MS2 data (and therefore rubbish Mascot results).

I've noticed recently that a high proportion ( ~80%) of the MS2 spectra consist of M- ~2 Da ions (compared to the precursor ion), and not a lot else. The precursors are usually designated as being 2+ and the products 1+, so I'm seeing (e.g.) 710 (2+) fragmenting into 708 (1+).

Has anyone else come across this? Any help would be gratefully appreciated!

Best wishes,
Lindsay.

I have never noticed anything like what you are descirbing (i.e. very unusual problem)... my suggestion would be (if you haven't done it already) to recalibrate all the parameters of your ion-trap (i.e. not only fragmentation but isolation etc...).

Also, just double checking:

What is the +2 m/z mass spectrum of your example; is it 355 or 356?

Are you using the recommended collision energies for peptide fragmentation?
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 1 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry