triacylglycerol analysis - poor resolution
Posted: Thu May 17, 2007 1:46 pm
In the past I have achieved excellent resolution of milk fat triacylglycerol peaks using a high temperature polarizable capillary column (Restek MXT-65TG, 30m long). However, the resolution has diminished after a number of major (installed a new FID jet and installed a new column - Restek Rtx-65TG, 30m) and minor (install, uninstall, reinstall, repeat 'til completely frustrated then try again) adustments. Ironically, it appears, that the more I tried to fix it, the worse my problem became. This led me to search the internet for clues to my problem - and that's how I found this forum.
In this forum, mtarrien (July 9, 2004) reported a similar problem, subsequently, on July 15, they reported that the issue had been resolved by using all new FID jets. Their conclusion was that graphite (from the ferrules used to hold the column in place) was accumulating on the internal surface of the FID jet and inhibiting elution or resolution of triacylglycerols in the sample.
In addition to replacing the FID jet I am considering making some other changes to my FID. Instead of installing a standard capillary jet I may install a high-temperature jet (suitable for simulated distillation). Since, high-temperature jets have larger orifices than standard capillary jets and may be less prone to graphite buildup. Additionally, I will most likely switch from graphite ferrules to aluminum ferrules (alumiseal ferrules from Restek).
I wonder if any of the readers of this board have similar or related experiences and can provide some guidance on these matters. Your input would be greatly appreciated.
Method details:
instrument: HP 6890 with split/splitless injector and capillary adaptable FID
column: Mxt-65TG and Rtx-65TG (Restek)
temperature program: injector and detector at 370C, oven temperature - 1 minute at 80C, 30C/min to 240C, 4C/min to 365C, held at 365C for 5 minutes.
Thank you,
John
Method details:
Thank you,