I run a standard 2 months ago and I measured an area on FID (with 60oC temp of the column. Yesterday I run the same standard with 100oC temp of the column and I measured much less area. Is this normal or the sample has decomposed?
-
- Posts: 397
- Joined: Fri Sep 22, 2006 7:36 am
Did you use freshly prepared solution(s) for your analysis yesterday?
Or is it the concentration of your analyte(s) very low, so the increase of the temperature can change the peak area little bit, due to the bigger noise at elevated temperatures? The integration of the small peaks could not ensure reproducible peak areas
Or you might have problem with your column, or the other critical parts of your GC system. Check for leaking!
Do you use headspace analysis or direct injection? If you're using headspace check your vials, whether thay are tightly closed or check your connecting parts on your transfer-line.
Also other reasons could be responsible why you're getting lower peak areas.
Supply us with further information for the method and instrumentation that you're using
Best regrads
Or is it the concentration of your analyte(s) very low, so the increase of the temperature can change the peak area little bit, due to the bigger noise at elevated temperatures? The integration of the small peaks could not ensure reproducible peak areas
Or you might have problem with your column, or the other critical parts of your GC system. Check for leaking!
Do you use headspace analysis or direct injection? If you're using headspace check your vials, whether thay are tightly closed or check your connecting parts on your transfer-line.
Also other reasons could be responsible why you're getting lower peak areas.
Supply us with further information for the method and instrumentation that you're using
Best regrads
-
- Posts: 12
- Joined: Thu May 12, 2005 5:06 pm
thanks zokitano for your response..
finally, I think the problem was the vial of my standard.. the septum was leaking..
finally, I think the problem was the vial of my standard.. the septum was leaking..
3 posts
Page 1 of 1
Who is online
In total there are 54 users online :: 0 registered, 0 hidden and 54 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 54 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 54 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science