I am trying to identify the 3 impurities on my compound that were nicely resolved in my method on LC. My question is, when those impurities went to the MS, will the resulted chromatogram in the MS would correlate to the LC part? Meaning, if impurity A eluted at 3.5, would i see a peak in the chromatogram in the MS screen at 3.5, too? Pls. enlighten and big thanks.
Albert Roy J
-
- Posts: 1074
- Joined: Mon Aug 30, 2004 5:42 pm
If your chromatographic conditions are the same then you will have almost the same retention factors (retention time may vary a little bit due to void volume differences).
Remember that you must use volatile mobile phase additives to be compatible with the MS.
Also, your compounds need to be ionized in order to be seen by your MS...
Remember that you must use volatile mobile phase additives to be compatible with the MS.
Also, your compounds need to be ionized in order to be seen by your MS...
2 posts
Page 1 of 1
Who is online
In total there are 355 users online :: 0 registered, 0 hidden and 355 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 355 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 355 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science