head space injection

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Dear all

I wanted to inject 500ul of sample in 1ml loop size. Is that possible. I heard that the sample size should be three times the loop to remove any dead volume.

please give me some information.

rajdeep

Normally you wish to sweep the loop of any residual carrier gas from the previous injection. (3 times the internal volume is recommended)

Why?

To inject consistent amounts of sample as most analytical methods which perform accurate measurements require it. If you do not require consistent amounts then you do not need to sweep the loop, but I believe that is very rarely done.

best wishes,

Rod

so how should i go about. I am using varian CP3800 with gas sampling valves, which can be attached to different loops.
3 posts Page 1 of 1

Who is online

In total there are 24 users online :: 3 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: deevencoo9, Majestic-12 [Bot], Semrush [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry