pulse-dampened pumping system for post column derivatization

Posted: Mon May 14, 2007 8:22 pm
by CQ_32
I'm looking for a pulse-dampened pumping system for post column reagent delivery. Then found the one as below :

http://www.waters.com/WatersDivision/Co ... 5&WT.svl=1

Any good experience with this one? There is not too many information on the website.

Thank you.

Posted: Mon May 14, 2007 9:17 pm
by Mark Tracy
This is actually an OEM from LabAlliance/SSI http://www.laballiance.com. It should work. The pulse damper requires back-pressure of ~1000psi in order to work correctly. Use a restrictor loop or a packed tube rather than a back-pressure regulator. You can also check Pickering Laboratories http://www.pickeringlabs.com if you want a deluxe system.