Here are a few more comments and clarifications from my previous post:
1) The choice of the value for a protocol acceptance limit will depend on parameters such as the type of analysis (assay, impurity, trace analysis, etc.) and the analysis technique (HPLC, CE, titration, etc.). Therefore, using different limits for different methods that are being validated is an acceptable, and common, practice. For the validation of an HPLC method for the assay analysis of a pharmaceutical drug product, a commonly used limit is +/- 2%. This is generally considered the obtainable accuracy for an HPLC assay method. Yes, HPLC can be better than that, but so far regulatory agencies have accepted the 2% limit and, although there have been pushes to tighten the limit, 2% is still common in this instance.
There should be consistency within and between validation protocols. If 2% is the limit for the solution stability in the protocol, then the protocol would also have 2% limits for the for the accuracy and precision experiments. Any SOP for method validation should contain typical values that are to be used (the typical values can be in and associated document such as a Work Instruction rather than in the SOP itself).
2) It is possible to store solutions at more than one storage condition but only test the solutions that have been stored at one only condition and not the other. The reason for this is that the second condition is a back-up storage condition. So, you keep solutions at ambient and chilled (i.e. refrigerated) but only test the solutions at ambient temp. Why? Well, you may not be sure that you solutions are stable at ambient but you don't want to repeat the experiment with a different storage solution. You would only do this if you suspect the solutions are not stable at ambient, otherwise you don't need the extra work. Whatever is done, the procedure for the experiment would be described in the method validation protocol. There would be a statement to the effect that the refrigerated solutions would only be tested if there were failures at the other storage condition.
3) There are two equations that can be used to calculate a relative difference:
A) (Tn - T0)/T0 X 100%
B) (Tn - T0)/((Tn + T0)/2) X 100%
Equation A would be used for a solution stability experiment as you need to compared the result for the stored solution to that of the initial (time zero) solution. For Equation B, you are making the calculation relative to the mean of two results. Thus, Equation B is NOT appropriate for the solution stability experiment. However, a variant of Equation B is appropriate for the precision experiment:
B) (R1 - R2)/((R1 + R2)/2) X 100% [variant]
In Equation B variant, R1 and R2 are the mean results for analyst 1 and analyst 2. As you are not sure which analyst has the true result, the equation uses the mean of the two results in the denominator.
When you have results that are around 100% (of label claim), then there is little difference in the calculated answer if either Equation A or B is used (i.e. the two equations give nearly the same answer). However, there is still a technical difference between the two equations even in the calculated results are nearly the same.
Note: For a method transfer, Equation A should be used with T0 being the result for the transferring analyst and Tn being the result for the receiving analyst.
4) Documentation, documentation, documentation! Keep your SOPs, Work Instructions and Method Validation Protocols complete. In your method validation protocol, state how you are to perform the experiment, what the acceptance limits are and how the results are to be calculated. Of course, your protocol should follow the procedures outlined in your SOP.
Regards,
Dan