We have 2 Agilent 6890 GC's. When we run our performance verification (6 injections of an FID test mix supplied by Agilent) on one GC, the peak areas are at about 400 pA. However, when we run the same sample on our other GC, the peak area is about 2000 pA. The method parameters are identical. This pattern is repeated across a number of our methods (methods include headspace and autosampler injecting, different columns, different liners, etc)with one GC response consistently about 20-25% of the other GC. The liners and grade of gases used on each GC are identical. Can anyone think of a reason? The only difference between the systems that I am aware of is that one system is capillary optimised, whereas the other has the option of using capillary or packed columns. On the system which has the loewr responses, the length of the jet for our FID is smaller than for the other system - the head of the jets and diameters are identical, it's just that one is longer by about 15 mm along it's shaft.