I have problems in establishing split ratio-splitting wanted relation.
At first I've menaged to have splitting measured with flowmeter around 40 ml/min when the split ratio set in the method was 40.
Nowadays at split ratio 40 I have value around 129-135 ml/min on the flowmeter.
What could be the cose?Why don't they follow each other?
Septum,column and detector are checked!
How to establish splliting arond 40 when I have set in the method split ratio at 40?????
Could possibly water in H2 geneartor that has lager concentration of co2 have influence?

Thank you!!!!!

The split ratio is the ratio of the flow out of the split to the flow into the column. Only if the flow into the column is 1 ml/min will the split flow be 40 ml/min at a split ratio of 40:1. At higher flows into the column the split flow will be higher for the same split ratio.

Peter
Peter Apps
Hi Peter!
Just want to thank you for the explanation.
Mail ya.
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry