CO2 determination with TCD

Discussions about GC and other "gas phase" separation techniques.

3 posts Page 1 of 1
Dear all,
I'm quite new in GC technique. My problem is that I need to measure pCO2 using a GC, with concentrations at least down to 300 ppm.
Our seller suggest us to buy a molecular sieve column (5A, 0.53 id, 30 m lenght), but I am not really convinced about this choice.
Does anyone know how this column work with CO2 and eventually suggest a better one?

Thank you in advance for all the help you will provide

best regards
Daniele

Your seller gave you very bad advice.

CO2 is absorbed on a MS 5A column. It is not recommended.

If you wish to use capillary then you need a porous polymer column like Supel Q made by my company Supelco.

There are other manufacturers of similar columns.

But if you want to measure low levels accurately then you should not use a capillary column but a micropacked column or at least a packed column as they will allow a much larger sample capacity than you achieve with a capillary column and will allow you lower levels of measurement with better accuracy. (It is also a lot cheaper than a capillary but money is no object, right? :D )

Supelco makes the best micropacked columns. I would suggest a 'R' or 'N' polymer for the best separation of CO2 from air and methane-ethane with a 80/100 or 100/120 mesh. 1-800-359-3041 for technical advice, or 1-800-247-6628 to place an order.

A 2 or 3 meter 1/16" OD 1mm ID column should be adequate for your analysis with a flow of 5 to 20 mL/min.

A sample of 100-500µL is suggested for use with a TCD or a helium ionization detector. With a methanizer and a FID you will be able to measure extremely low levels, far below your probable needs, but this requires hydrogen carrier. Helium carrier is suggested with a TCD.

best wishes,

Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823

814-359-5737 voice
814-359-5459 fax
rodney.george@sial.com

Dear chromatographer1,
thank you very much for your fast reply. you confirmed my suspect, even if the seller continue to say that the column is ok to detect CO2, I will ask to another one.

I will also look for the column you suggest. Maybe in the next days I will ask you some other suggestions.

best regards
daniele
3 posts Page 1 of 1

Who is online

In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry