- Posts: 13
- Joined: Thu Apr 12, 2007 12:41 pm
Now, I have a big problem with my new column, I only can see the solvent peak, when i injected the standards(there should be 4-37 compounds), what is the major reason?
thanks a lot for your support!
Elaine
Advertisement
Discussions about GC and other "gas phase" separation techniques.
Yes, but in this case she doesn't get any peak from her standard. The column length and film thickness do not affect whether the peaks will come out or stay in the column. The lenght affects the time of retardation of the components in the column. However she waited and waited but the peaks do not come out. So the main problem is from other nature. The change of the parameters of the column I think wouldn't show this kind of problem.One possible reason is that you are using the same parameters as for the old column. You need to adjust the oven temperature program as well as inlet pressure when a 30m column is replaced with a 60 m column. Because film thickness is 0.15 um for 60 m column, you need to inject less to avoid overloading.