Fingerprinting by GC-FID

Posted: Mon May 07, 2007 10:16 am
by RAMU
has anybody tried to use the FID to fingerprint ( to produce some kind of library based on retention time of the standard)

Posted: Mon May 07, 2007 1:05 pm
by chromatographer1
The answer is YES. Used in hydrocarbon analysis for decades.

best wishes,

Rod

Posted: Mon May 07, 2007 2:35 pm
by philippem
Kovas indices were/are used for doing such kind of fingerprinting. This deals with the selectivity and retention on your statinoary phases.
FID is used as detector ( all purpose detector, as long as there are enough carbonatoms in your compound) in order to "see' your compounds of interest.

succes

Philippe

HYDROCARBON FINGERPRINTING

Posted: Mon May 07, 2007 6:13 pm
by RAMU
CAN ANYBODY SUGGEST SOME BOOK THAT EXPLAINED THE CHROMS IN DETAIL, COMPARED IT TO THE EPH standard.

Posted: Tue May 08, 2007 6:51 am
by philippem
RAMU,

please check the FAQ page , can be found on the left side of the main screen.

here you will find info about chromatography and further references

success

Philippe

Posted: Wed May 09, 2007 2:25 pm
by AICMM
Ramu,

Kovats indices are listed for many compounds on the NIST webbook web page. Furthermore, the ATF (Alcohol, Tobacco and Firearms) has an excellent arson investigation book that compares some 80 or so flammable mixtures against a normal alkane standard mix run under identical conditions. Unfortunately, I do not know how to obtain a copy (there is nothing on my version that tells me how) but if you are looking at hydrocarbon mixtures it would be worth the effort.

The ATF and the Kovats both require you to run a normal alkane mixture to establish retention times that can then be compared to the reference.

Best regards.