TCD for N2 purity

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Dear all
which TCD do you suggest me to measure from 99,5% to 100% N2 if I want to have an accuracy and repeatability on 12 hours of 0,01%?
thanks

almost anyone's TCD should work.

The issue is injecting a reproducible amount of N2

which does not overload the column, not the TCD per se.

The integration capabilities is quite important as well.

Usually folks measure the impurities in a balance like N2 and subtract their values from 100% to get a measurement of N2 purity at ±0.01% reproducibility.

Don't try to measure the balance peak with this precision as it is not possible due to the changes in atmospheric pressure which cause a variation larger than 0.01%.

best wishes,

Rod
Rod
We presently use a differential TCD which compare pure Helium at 30cc/min with the N2 sampling volume at 30cc/min. Sampling loop size is 2ml.
We use a molecular sieve 5A 8' heated at 80 Celcius and we get a integration time from 50seconds(peak starting time) to 90 seconds(peak ending time).
We calibrate the span with pure N2 at 100% and the zero with pure Helium. When we run cycles on pure N2 we have a variation around +/- 0,2% on 2 hours period. Is there a way to have better repeatability??

It would appear that you are getting normal performance from a sampling valve system from a laboratory GC.

2 out of 1000 repeatability is not bad.

I don't think there is any way of improving that unless you go with a very highly designed system, a special process analyzer.

Measuring a balance peak directly is not the way to get good repeatability on that peak.

You must measure by difference between 100 % and the sum of the impurities.

I repeat again:

You must measure nitrogen by difference between 100 % and the sum of its impurities.

If I have not made myself clear to you please give me a call or write me privately and I will attempt to clarify my answer.

best wishes,

Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823

814-359-5737 voice
814-359-5459 fax
rodney.george@sial.com
4 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry