has anybody tried to use the FID to fingerprint ( to produce some kind of library based on retention time of the standard)
-
- Posts: 3210
- Joined: Thu Sep 02, 2004 7:28 pm
The answer is YES. Used in hydrocarbon analysis for decades.
best wishes,
Rod
best wishes,
Rod
-
- Posts: 71
- Joined: Tue Jul 05, 2005 6:23 am
Kovas indices were/are used for doing such kind of fingerprinting. This deals with the selectivity and retention on your statinoary phases.
FID is used as detector ( all purpose detector, as long as there are enough carbonatoms in your compound) in order to "see' your compounds of interest.
succes
Philippe
FID is used as detector ( all purpose detector, as long as there are enough carbonatoms in your compound) in order to "see' your compounds of interest.
succes
Philippe
-
- Posts: 2
- Joined: Mon May 07, 2007 9:50 am
CAN ANYBODY SUGGEST SOME BOOK THAT EXPLAINED THE CHROMS IN DETAIL, COMPARED IT TO THE EPH standard.
-
- Posts: 71
- Joined: Tue Jul 05, 2005 6:23 am
RAMU,
please check the FAQ page , can be found on the left side of the main screen.
here you will find info about chromatography and further references
success
Philippe
please check the FAQ page , can be found on the left side of the main screen.
here you will find info about chromatography and further references
success
Philippe
-
- Posts: 1773
- Joined: Fri Feb 09, 2007 10:33 pm
Ramu,
Kovats indices are listed for many compounds on the NIST webbook web page. Furthermore, the ATF (Alcohol, Tobacco and Firearms) has an excellent arson investigation book that compares some 80 or so flammable mixtures against a normal alkane standard mix run under identical conditions. Unfortunately, I do not know how to obtain a copy (there is nothing on my version that tells me how) but if you are looking at hydrocarbon mixtures it would be worth the effort.
The ATF and the Kovats both require you to run a normal alkane mixture to establish retention times that can then be compared to the reference.
Best regards.
Kovats indices are listed for many compounds on the NIST webbook web page. Furthermore, the ATF (Alcohol, Tobacco and Firearms) has an excellent arson investigation book that compares some 80 or so flammable mixtures against a normal alkane standard mix run under identical conditions. Unfortunately, I do not know how to obtain a copy (there is nothing on my version that tells me how) but if you are looking at hydrocarbon mixtures it would be worth the effort.
The ATF and the Kovats both require you to run a normal alkane mixture to establish retention times that can then be compared to the reference.
Best regards.
6 posts
Page 1 of 1
Who is online
In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 24 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 24 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science