- Posts: 10
- Joined: Wed Apr 25, 2007 2:28 pm
I´m currently working on a method development, I must analyze dehydrocholic acid in tablets, but I have two main problems:
1 - I work mainly with HPLC/UV, and the dehydrocholic acid does not absorb in the working UV range.
2 - The matrix includes an artichoke extract, which gives me a lot of unwanted peaks on the chromatogram...
What I´ve been trying so far is the derivatization of the dehydrocholic acid with DNPH, in the following way:
1 - The sample is dissolved in MeCN:H2O (90:10) (about 0.04 mg of DHC per ml)
2 - The DNPH reagent is prepared dissolving 2 g of DNPH in 2 ml of H2SO4, completing the volume (100 ml) with MeCN:H2O (90:10)
3 - The derivatization procedure is: 4 ml of Sample + 1 ml of DNPH reagent, 120 min. at 60ºC, completing the volume with MeOH:H2O:AcN (83:2:15) (so the final proportion of solvents equals the mobile phase)
I am using a Kromasil C18, 25 cm, and I can integrate a stable peak, sensitive and linear to the concentration changes, but I can see a lot of other peaks that keep on growing with the time. I think that my derivatization procedure is failing somewhere, but I´m not used to work with DNPH... can anyone help me with this, or at least sugest any other way to analyze the DHC???
Thank you all...