Missing Peaks
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
3 posts
Page 1 of 1
I changed the outlet tubing of autosampler with new one and all peaks were missing though the flow was ok. Could a new tubing absorbed the analytes?
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
I doubt it, my guess is some kind of leak. One question is why did you change the tubing, and second is did you change anything else there as well. If neither of those, inject some 10x concentration and see what happens then.
-
- Posts: 480
- Joined: Fri Sep 03, 2004 7:57 am
... the sample loop was disconnected by someone or 'something'. problem has been solved.

3 posts
Page 1 of 1
Who is online
In total there are 552 users online :: 1 registered, 0 hidden and 551 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 551 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Semrush [Bot] and 551 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science