m/z 254 and m/z 300-500 contamination

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
I found out the contamination ion intensify at m/z 254 and many ion between m/z 300 - 500. These can be observed even no injection. I used H2O+Formic acid+ammonium acetate and Acetonitrile for mobile phase.

I changed mobile phase to only MilliQ water and Acetonitrile, however, these ions still can be observed. I used gradient mode. These ions are always observed in period of acetonitrile.

Are there anyone had the same experience or any suggestion to clear up these contaminant?

Thank you,
Areeya

What grade solvent are you using. We have found that it is necessary to use a very high grade solvent to avoid seeing these kinds of problems. A bit more expensive, but should solve your problem. Alternatively, you can distill your solvents prior to use (preferably multiple distillations). Or just try chromasolv from Sigma-Aldrich:

Part number= 34967-2.5L
2 posts Page 1 of 1

Who is online

In total there are 35 users online :: 1 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry