That's a definite "maybe". A simple "mass detector" (usually a single quadrupole instrument) should work, but there are several potential problems:Is this a potential LC/MS application or am I best advised to try another tack ?
- if your "typical reverse phase C18 setup" uses something like a phosphate buffer, you will have to re-work the separation (non-volatile buffers are a problem; some of the newer systems will tolerate them for short periods of time, but they are best avoided)
- a very complex sample like an herbal extract have other junk with an m/z of 441 eluting near your folic acid (and I'm not enough of a mass spectrometrist to tell how well folic acid will be ionized and detected).
Before spending a lot of time or money, I would talk to your friendly local LC-MS sales person (from Thermo, Waters, Agilent, Shimadzu, and whomever else I'm missing) and see if they would run a test sample for you to evaluate feasibility.
You can almost certainly do it via LC-MS/MS using a triple quadrupole system. These are routinely used for bioanalytical work looking at very low levels of drugs or metabolites in biological fluids or tissues. You end up looking at specific fragments of specific molecules, so you can deal with the very complex samples. The catch is, they ain't cheap (you're looking at >$200k).