In most industries, HPLCs are used in three stages.
1. Raw materials - Did I get what I might get around to paying for?.
I assume stock control is OK ( big assumption, "we need release tomorrow, and the sample arrive in the morning" and generally the sample will actually arrive at 1600, if lucky... ). The HPLC method is only a small part of the quality checks required to release the product, so run times usually aren't too important provided stock control is good.
These are usually well defined, often have to have gradient method because the precusors and contaminants can be quite diverse, For these gradients ( which can cover quite a range of mobile phase polarity ), I like to have a cycle time of 45 mins +-15 mins, and normally default use is a C18 column, but often special-purpose columns and conditions are used. They are usually optimied for resolution and sensitivity.
2. In process controls - is it ready for the next step yet?.
These are inevitably time-constrained. 10 mins cycle time is desired, and usually the precusors and products are fairly similar, so even isocratic can be used. Generally very shallow and short gradients, just to keep the cycle time down - I'm a big fan of the old wives' tale - maxiumum 10%/min composition change.
The method must be availble on demand, so cycle time is critical, and customers may only want to see the changes in one peak, and ignore the rest. Typically, I use a 150mm x 4.6 with guard, and target retention of 6-8 minutes, with last peak at 9.5 minutes. If people want more resolution of minor peaks, a 250mm column is used, and they accept the additional time.
One critical choice about these analyses is the HPLC column robustness, in general, a C8 column will croak if fed with junk, whereas a C18 may survive. I pay quite a bit of attention to column durability, because you really want to minimise any sample preparation for IPC. Because of this, sometimes a 4.6mm ID is preferred, just because the guard copes with more junk before dying. Often the junk isn't visible at the wavelength used, or even visible at all - I may choose a non-optimum wavelength to lose interfering junk, rather than optimise the method.
In some IPC cases, I have to consider the sample matrix, and choose column/mobile phases that are compatible. The simplest and fastest may be the best, even if guards only last a few runs. Internal standards are seldom used, as sample baselines often look like the Himalayas
3. Final Product
Have we met or exceeded the product specification?.
Often this will also be the raw material for client, so the methods tend to be tailored to that. Long gradient or isocratic ( isocratic run time is 3x actual peak retention time ), sometimes with differing methods looking for precusors, related substances, degradation products, but often a wide gradient of 45+-15mins keeps everybody happy.
These are similar to compendial methods, with all the associated nausea. I tend to be able to choose any column that provides reasonable separation, and column robustment and method sensitivity ( <0.1% ) usually aren't issues,
These are well defined products, so behaviour in acid/base mobile phases, temperature etc, are all checked out, as well as ensuring all peaks of interest ( usually to 0.05% ) are resolved. I don't want samples to degrade during testing. If all peaks aren't resolved, then another method may be added ( eg related substances ) to address the need. These will have the most method refinement, even without any cGMP stuff.
Same sort of instrument timeline as for raw materials ( 45+-15 minutes ), as some other tests usually are required on such samples. These methods are the most optimised for resolution, accuracy, and repeatability.
Obviously, if labs are instrument limited, then faster methods are only advantageous if similar systems and columns are used, as you don't want to keep swapping. However the ability to not croak with real life samples is important, as is minimal sample preparation.
In most methods, samples would be dissolved in mobile phase, and playing with column temperature, mobile pahse pH, etc is constrained unless I know the sample is stable. Rejecting product because of an analytical artifact isn't considered as agood deed by most managers
Note that places that setup to run studies, eg clinical trials, or specific analyses ( environmental ) will only want methods that look for one or two peaks, and they want even faster cycle times than IPC. That's usually 3-5 minutes, and with MS. Method development may involve an Internal Standard for those.
I hope this is helpful,
Bruce Hamilton