First off, on column degradation (or on column transformation i.e. from one ionic species to another, much more common than degradation) is relativly rare, although it does happen. It usually shows itself as a very distictive peak shape, often described as a shelf or bridge type of peak. What it looks line is a sharp upward rise, sharp down to a point that is much higher than the baseline level and then off this shelp a fairly normal looking peak. So, what you have is the parent on one end, the product on the other, and the molecules that are in transition forming the shelf or bridge. You can often use temperature to drive the reaction one way or the other to confirm if this is what is really going on. If your sample is truly degrading, it is more likely happening during the sample workup, cleanup or what ever other manipulations you do to it befor injection. I dont think I have ever heard of salts in a sample causing this, most salts that are in a sample just elute in the void (in reverse phase). This is why people like to use SPE to de-salt samples. The salts are unretained on the SPE cartridge and go to waste, and the compounds of interest can then be eluted.