I have a problem with peak splitting and I'm hoping I can get some ideas.
The method is a 0.1% formic acid / ACN gradient from 5-95%B on a PackPro C18 column. Sample concentration is 0.2 mcg/mL with 20 mcL injection volume. Sample solvent is 50/50 ACN/Water.
This method has been in place and running well for ~18 months when suddenly, I observe peak splitting of multiple components.
I switched columns (three different) and systems (N=2) and observe the same problem, though it does vary in intensity between columns (from shoulders to nearly resolved split peaks).
All mobile phases, diluents, and samples have been reprepared with no significant changes.
Preparing samples in weaker solvent or decreasing the injection volume does seem to improve peak shape. So the obvious solution is to do just this.
However, this is a validated method and it has been working well for all this time. What can be causing this sudden change?
Thanks for taking the time to consider this problem and I appreciate any ideas.