SEC/GPC Standard Issues
Posted: Thu Apr 19, 2007 4:07 pm
Hi everyone!
I am developing an SEC method on a Phenomenex BioSep 2000 column 300 x 7.8mm using 1xPBS as the mobile phase and 10% DMSO to remove any residual proteins. The flow rate is 0.3 mL/min. This method is run on a Waters Alliance 2695 system with UV detector at 280 nm. This method will be qualified, transferred to a contract manufacturing facility and validated eventually. From my previous life, I was under the impression that a linear curve and system suitability most be run for each assay. I purchased SEC calibration standards from BioRad containing
Thyroglobulin 670,000 Dalton
Bovine gamma-globulin 158,000 Dalton
Chicken ovalbumin 44,000 Dalton
Equine myoglobin 17,000 Dalton
Vitamin B12 1,350 Dalton
When I set up the calibration curve using GPC software from Empower I got a MW of my protein as 270 kD when it is really 64 kD. My protein elutes at 21 minutes and which came up between the first two standards. My protein is a fusion protein, so I was wondering if it is too different from the standards and that is why I am getting a higher MW? The column I use has a molecular weight cutoff of 300 kD, would this be the problem? It is a requirement that a linear curve be run from a regulatory standpoint?
I appreciate the help! Have a nice day!