I have one Agilent LC/MSD Trap XCT Plus. Recently the sensitivity is pretty poor (in regular ESI mode), for most compounds, the lowest concentration I can reach is ~100ng/ml or even higher. It should work at least ~1ng/ml. Any ideas to result in this? I clean the ESI glass capillary (it looks like not dirty when I take it out), no improvement. Not sure if there are some hardware problems involved. Thanks.
-
- Posts: 1074
- Joined: Mon Aug 30, 2004 5:42 pm
Maybe your electron multiplier is dying?
I recently clean ours until the octapole but it seems that there is no much passing the first orifice before the octapole (it looked pretty clean). The orrifice itself before the octapole was not clean...
How about the vaccum? About the same as always or a little bit higher vaccum (i.e. lower pressure)?
I recently clean ours until the octapole but it seems that there is no much passing the first orifice before the octapole (it looked pretty clean). The orrifice itself before the octapole was not clean...
How about the vaccum? About the same as always or a little bit higher vaccum (i.e. lower pressure)?
-
- Posts: 1
- Joined: Sat Apr 14, 2007 1:42 pm
Are you sure that the trap itself and the ion optics are clean? e.g. when you have a power loss and the ballast port on your rough pump is open oil can be sucked back into the system. Therefore keep it fully closed all the time and open it every two weeks for a few minutes. Beside that the nebulizer has a huge effect. Check it with a magnifier.
Regards,
Bykling
Regards,
Bykling
3 posts
Page 1 of 1
Who is online
In total there are 20 users online :: 0 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 20 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 20 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science