2 posts
Page 1 of 1
I'm trying to enter standard weights, purity corrections, and dilutions separately into Empower software (instead of pre-calculating a corrected concentration in mg/mL) and get a two-point linearity curve (linear thru zero) with the same equation as when I enter the pre-calculated concentrations in mg/mL. Is there anyone out there who has any experience with a similar situation?
-
- Posts: 143
- Joined: Mon Aug 30, 2004 8:10 am
You should make a custom field which calculates the amount for that peak, and use this custom field to calibrate with, instead of amount.
2 posts
Page 1 of 1
Who is online
In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Google [Bot] and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science