neuger » Wed Apr 11, 2007 2:16 pm
can anyone explain this? dozens of injections in neat sol'n can produce fairly consistent retention times with a gradient of 2% - 14% ACN (across 6 min) with 0.1%TFA being the other mobile phase, yet the moment simple precipitation w/ 4 parts MeOH to 1 part plasma, evaporate supernatant and reconstitute with MP and inject, the retention moves almost 2 minutes at times and doesn't even really stabilize. This is such a simple precip method, I just don't get it!! Can such simple solvents like MeOH change the chemical makeup when they are used for precipitation of proteins, never seen that before? Could evaporation and concentration be a problem, again never seen it be a problem? If so, that is bad for me, I need it to achieve the sensitivity for the clinical samples we are testing.
I am pretty sick of this happening...here are the particulars:
YMC-C18 hydrosphere @ 30 degrees.
A: 0.1% TFA in H20. B: ACN w/0.1% TFA (I have tried with and w/o the TFA in ACN, it doesn't matter!)
reconstitute w/ 90:10 (A/B), inj 50ul.
retention in neat sol'n = approx 9.1 min
retention after reconstitution of supernatant = shifts from 9 - 8.5 - 7.1, etc
a large wash out step has been added to 40% organic at times and the run has been as long as 40 min to attempt to acheive equilibration, but none of this has helped...and you all thought I wasn't washing and re-equilibrating after a plasma matrix...lol, but seriously, anyone think they know what's up here??