It looks like your problems could be both hardware and assay protocol.
That Waters pump is rated for 5 ml/min , if the pump is complaining - then you could have mechanical problems.
Just to guess - are you starving the pump inlet with clogged solvent filters or are the buffer reservoirs below the level of the pump?
Concerning your assay -Step bumps are recomended by Applied Biosystems and also every procedure I have seen using this type of column.
I see no reason to use a gradient elution, especially with phosphate buffers!
When you are between pH 3 and 5 you have NO BUFFERING capacity and your eluted protein is probably the strongest buffer.
Think of the theory of what would this do to the peak shape???
A guide for this column can be found at :
http://docs.appliedbiosystems.com/pebio ... 003401.pdf
This discusses a problem with "refractive index changes with phosphate elution buffer" causing distorted peaks with your low concentration standards and blanks.
I also have some other literature on validation of an assay for MAb's using this column. Unfortunatly I can not find a current web link for the paper. Send me your email and I will forward it to you.
I appoligize if I sound critical, it is just that you have hit on a topic that I have spent too many years working on.
I do understand how frustrating it is when you are given a protocol that is "set in stone".
If you can't reproduce it ... it makes you look bad.
Also this might be a low prority for your time, you just have to get this done and move on to something else.
It is one of my faults that I have a low tolerance for "Bad Science" repeated forever , because "that is how it has always been done" !!
(Been there - Done that .... I have not always made friends in the process.)
Good Luck and Have FUN !