Still confused about Agilent 6850 flows and pressures
Posted: Fri Apr 13, 2007 3:53 pm
So after getting input from here, I set one of my GC's to a constant column flow of 1.5 mL/min (constant flow), with the "Purge Flow to split vent" = 0 mL/min and the GasSaver option OFF - thus yielding a total flow of 4.4 mL/min (splitless mode). The oven, electrometer, nitrogen, air and hydrogen are all set to OFF (this is my standby mode).
The problem came when I tried to apply this same method to my second GC (also an Agilent 6850). The inlet pressure does not maintain the setpoint - it keeps going down steadily, eventually causing the GC to shutdown. This GC does have a different column installed and I thought that perhaps the flow was too low for the column in use. But when I raised the column flow to a higher value, the setpoint could still not be maintained. The only ways that I found I could keep a steady inlet pressure was by either turning on the GasSaver feature (minimum of 15 mL/min - thus bringing the total flow to 19.4 mL/min) or by setting the "Purge Flow to split vent 2.00 @ min" to at least 1.0 mL/min (giving a total flow of 5.4 mL/min when column flow = 1.5 mL/min).
Furthermore, I found that after a week, the first GC started the same thing and could not maintain the inlet pressure and I thus set the Purge Flow to split vent 2.00 @ min" to 0.5 mL/min (giving a total flow of 4.8 mL/min when column flow = 1.5 mL/min).
Can anyone explain this to me...?
Thanks!
Roxanne.
The problem came when I tried to apply this same method to my second GC (also an Agilent 6850). The inlet pressure does not maintain the setpoint - it keeps going down steadily, eventually causing the GC to shutdown. This GC does have a different column installed and I thought that perhaps the flow was too low for the column in use. But when I raised the column flow to a higher value, the setpoint could still not be maintained. The only ways that I found I could keep a steady inlet pressure was by either turning on the GasSaver feature (minimum of 15 mL/min - thus bringing the total flow to 19.4 mL/min) or by setting the "Purge Flow to split vent 2.00 @ min" to at least 1.0 mL/min (giving a total flow of 5.4 mL/min when column flow = 1.5 mL/min).
Furthermore, I found that after a week, the first GC started the same thing and could not maintain the inlet pressure and I thus set the Purge Flow to split vent 2.00 @ min" to 0.5 mL/min (giving a total flow of 4.8 mL/min when column flow = 1.5 mL/min).
Can anyone explain this to me...?
Thanks!
Roxanne.