hi
I would like to know which detector is the best if I want to measure 1ppm Helium in Nitrogen. Which detector is sensitive enough.

thank you

Tigerk,

About the only detector that can measure helium at 1 ppm is a TCD and you should be able to measure at that level if you use nitrogen as the carrier gas. Nitrogen should be five or six nines purity or as pure as possible as far as helium is concerned.

I would use a 3 to 6 meter Carboxen 1000 packed column, 1/16" OD 1.0 or 1.25 mm ID , a moderately slow flow rate and use at least a 1mL loop at ambient temperatures if possible. The only probable impurity that might give you a false reading is Neon.

best wishes,

Rod

what will be the LDL if I used pure N2 as carrier gas for the measurement of Helium?

My guess would be with a TCD the limit would be in the range of 1-10ppm, even less depending upon a lot of factors. I recall measuring 1-10 ppm of hydrogen in an application about 20 years ago. Helium's LDL should be similar.

best wishes,

Rod
4 posts Page 1 of 1

Who is online

In total there are 85 users online :: 1 registered, 0 hidden and 84 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 84 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry