Hollow » Wed Mar 21, 2007 7:05 pm
Thank you all for your replies
After the few tests we made on both systems I can "agree" with Mattias.
On the RRLC we had much less problems to run an adapted HPLC method.
On the UPLC, we get some more troubles. Unfortunately we couldn't figure out what was really the problem.
The problem was that we often get "split" or unsymmetrical peaks in the optimum flow rate range (400-600µl/min on a 2.1x50 mm C18 UPLC column).
Once it was caused by the sample solvent/injection volume, but even with a alkylphenone test mix it was there.
I've run a sequence with different flow rates, starting with 100ul/min up to finally 700ul/min of isocratic MeOH/H2O 40/60 (or was it 60/40? -> similar to Waters PQ test for UV-HPLC system).
As said, between 400 and 700ul/min the last two peaks (propio- and butyrophenon) gets split. Maximum split was at 500ul/min.
The phenomenon was reproducible, a 20x injection gets %RSDs of 0.1 in retention time and about 1.3 for (combined) area.
Interestingly the peak form gets better again with a flow rate of 700ul/min.
The same thing was about to be seen on a 2.1x100 mm column, but here I had to raise the temperature from 40 to 70°C to get a similar flow rate of 500ul/min. Higher rates than 570ul/min weren't possible cause of the pressure limit.
We told this to Waters CH, but they haven't got a solution yet (okay, it was just last week...)
@Uwe: Have you got any idea what it could had been? If you're interested in further details I can send them by mail.
Regards