Internal standard problems with EPA Method 524.2
Posted: Tue Apr 10, 2007 7:59 pm
We are analyzing VOC's in water using GC/MS. We have noticed a pattern in which the response of our internal standard fluorobenzene is increasing as the concentration of our calibration standards increases. We did notice that the response of the flourobenzene was constant between replicates of the same calibration standards so we assume that the correct amount of internal standard is being injected by the autosampler.
We realized this problem when trying to run quality control samples from an outside source and all the values were consistently higher than the actual known values. When we took out the internal standard and re-evaluated the same run, the quality control results then passed.
Does anybody know what could be causing the internal standard to increase. We did notice that between two different sets of quality control samples containing different compounds, that the flourobenzene response was consistently higher in one and lower in the other. This leads us to suspect there are interfering ions or we also read about active sites. Could either of these be our problem?
Instrument Conditions:
Column: RtxVMS
30m 0.25mm 1.4um
1ppb internal standard per 25mL sample purge
Helium carrier gas
split ratio 30:1
Thank you so much,
Column: RtxVMS
30m 0.25mm 1.4um
Helium carrier gas
split ratio 30:1
Thank you so much,