- Posts: 87
- Joined: Fri Mar 24, 2006 10:14 pm
We have an Agilent GC/MS 6890N/5975 system. I guess because our system is dirty. The sensitivity is really bad now. I used OFN 1pg/ul to check sensitivity. It is getting worse and worse. S/N is not that bad. I got 162, 66 and 56. But Signal-Avg nois=1443, 617, 423, which was 10000 a year ago. So my LOD is bad now.
I did air/water check
18/69=0.9%
28/69=0.25%
32/69=0.15%
44/69=0.09%
28/18=27.55%
If I compare tune history files, 502 relative abundance is lower (was 19 now 9); EMVolts is higher (was 1247 now 1718); IonFcus is lower (was 90.2 now 76.9), Entlens is lower (was 25.5 now19), ENtoffs is lower (was 20 now18) and Repeller is lower (was 27 now25.77).
I used Aero duster (peak 33) to check air leak, and did not find any.
In manual tune mode, the abundance for 18 is large (40000). The abundance for 28 (3466) and 32 (1629) are fine.
Recently we injected water solutions and large amount of dynamite samples.
What can I do to improve my sensitivity? Do I need to clean the ion source?
Thanks a lot!