Pegry » Thu Apr 05, 2007 6:28 pm
Hi Davide
Gasman is probably right, and there are some other possibilities.
When you say the "last point of the calibration curve" is that the highest or lowest concentration of analytes ?.
How are you adding the internal standard to the calibration standards ?, at what stage and by what technique ?
What is the rsd of the analyte peak areas for the replicates of the last point ?
What shape is the curve of analtye area vs concentration, curved ?, and what is the r-squared ?
What is the shape of the analyte/i.s. vs concentration, and the r-squared ?
If you do a blank injection after a standard, do you see any caryyover ?
Peter
Dear Peter,
I've tried to inject six times both the highest and the lower calibration point but i got always a low RSD (about 3%). I add the internal standard by volumetric pipetting in each vial the same volume of internal standard solution (i.e. 100uL).
The replicates of the area of the analytes are similar to the internal standard (about 3%). The calibration line (analyte/i.s. vs. concentration) is linear with a r-squared of 0.999. I always inject a blank after the higher standard but there isn't carry over.
Is the first time that I face such a reproducibility problem but maybe is related to the poor evaporation of the tryglyceride in the injector? I have to add that before every calibration we do a system suitability with the middle concentration standard and we inject that standard six times with good rsd (<3%).
The strange thing is that the area of the internal standard increase after each injection when doing the calibration but remain constant when doing system suitability...
If you have some cue please tell me because I would like to know how to solve this problem..
Thanks in advance and have a good day