Isolation in ion trap MS
Posted: Thu Apr 05, 2007 8:09 am
				
				WE face a problem using Ion Trap DECA XP Thermo
We found in the same analytical run that some molecules can be isolated for MS-MS experiment, while some others can not.
This fact happens only in negative acquisition mode. The same compounds can be isolated at 100% in positive
For example we can isolate (isolation width 1 amu) estradiol, estrone but not ibuprofen and ketoprofen (which are benzyl-acetic acids). Also Ultramark ions can not be isolated in negative (while can be isolated in positive)
We can recover signals if we apply a isolation width of more than 4-5 amu.
We want to know whether:
- It depends on the kind of molecules (and that case the Ion trap is not suitable to analyse these compounds and nobody told us)
or it is a problem of our instrument
			We found in the same analytical run that some molecules can be isolated for MS-MS experiment, while some others can not.
This fact happens only in negative acquisition mode. The same compounds can be isolated at 100% in positive
For example we can isolate (isolation width 1 amu) estradiol, estrone but not ibuprofen and ketoprofen (which are benzyl-acetic acids). Also Ultramark ions can not be isolated in negative (while can be isolated in positive)
We can recover signals if we apply a isolation width of more than 4-5 amu.
We want to know whether:
- It depends on the kind of molecules (and that case the Ion trap is not suitable to analyse these compounds and nobody told us)
or it is a problem of our instrument