API 2000 X OPOSITE OF ION SUPPRESSION EFFECT

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
:cry: :cry: :cry: :cry: :cry: :cry: :cry: :cry: :cry: :cry: :cry:
Dear Friends

We are facing a very strange problem.
We are analyzinf plasma samples (100 ul) extracted with hexane. The mobile phase is acetonitrile - 0,05% formic acid (55:45) - 10% isopropanol. After the extraction, the residue is evaporated and ressuspended in 200 ul of mobile phase. The injection volume is 8 ul.
The column is an X-Bridge C18 (Waters). The run time is 3.2 min anf the flow is 0,6 ml/min (with split of 1:2)

All this study was done 6 months ago, with no problem at all. We are doing another study of the sampe analyte for another client.

At this study, with other plasma samples, also from volunteers, we have a kind of contamination, when, after 20 or 30 samples, the analyte area increases.

We change all the autosampler injector parts as needle, needle seat, rotor seal, peek tubes, and clean all the MS system.

We also introduced a wash step of the needle in the injector and a injection of a mobile phase in a different method (100% of organic for 4 min and 100% of mobile phase for 2 min) every 8 samples.

Nothing of my tests showed me the answer, in other words, the problem is still there.

Does anybody have this experience or could give me an idea ????????????

Thank you.

Doesn't have to be suppression, can be enhancement. See page 13 reference 7 on my webpage about enhancement due to phopholipids. If not eluted, the lipids will start to break through and can give enhancement or suppression depending the elution time of eluent..

http://users.chartertn.net/slittle/ see the matrix effects in LCMS section.
Sailor

It depends on the analyte, but we also see very pronounced ion enhancements with certain analytes.
Matthew

I agree. Phospholipids are a likely cause of the problem. I would wash the column after the twenty or so samples with 100% organic to get rid of the non-polar contaminants. Alternatively, an improvement in the sample preparation step would solve the problem.

If you are going to wash, probably need to wash everytime a sample is injected or at least every 3-4 samples. After that, the phospholipids start to break through and affect analyte response.

However, does seem to reach steady state and doesn't change after a few injections.

See my paper at my website or in J. Chromatography

http://users.chartertn.net/slittle/
Sailor
5 posts Page 1 of 1

Who is online

In total there are 23 users online :: 2 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry