I have a situation in which the back pressure is increasing with each injection. I am using 60% (25mM KH2PO4 buffer, pH 6.3) w/ 40% ACN on a C18 50x4.6, 3.5µ. Can anyone give me some help on identifying what the issue is and how to correct? :cry:

Thanks

There is no reason that your mobile phase should have a solubility problem. Most likely you have particulate matter in your samples. Filter them; if you already did, use a smaller pore-size filter. You didn't say what the samples are, but soluble or colloidal proteins and carbohydrates can precipitate on contact with the mobile phase; check this by a test-tube experiment. Possibly your mobile phase needs filtering. Less commonly, a filter element inside the pump may be needing a change.

Once you find the source, it will be a lot easier to decide how to treat the column.
Mark Tracy
Senior Chemist
Dionex Corp.

Try to premix the mobile phase (if you havn't done that already)

Some instruments cannot do this mix without precipitation. See Mark's comments otherwise.
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 0 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry