Dear All,

We are having difficulties finding a method for the detection of DMSO.
We have tried headspace and also liquid injections.

Can any one provide us with a method that can quantitate 80 ppm?

thank
DMSO is a sticky problem.

The matrix can make the low detection levels for DMSO difficult if not impossible. This affects the choice of solvent and the side reactions which can take place. Also the solvent affects how well you can refocus the DMSO during injection.

LC may be a better solution for you.

SPME is another option to try.

Suggestions for GC analysis are:

injector temperature high enough to volatize the DMSO well

proper choice of column which will not react with the DMSO and refocus effectively in the matrix involved

Wash solvent plug behind the sample plug in the injection syringe

Clean injection liner is a likely requirement (replacement may be needed after each sample injection)

good luck and best wishes,

Rod
2 posts Page 1 of 1

Who is online

In total there are 72 users online :: 0 registered, 0 hidden and 72 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 72 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry