I don't know guys. Is it really that complicated?.
Many of those mechanisms are used by the Vancomycin column that I'm current trying to achieve a chiral separation on. The screening mobile phases are:-
Polar Ionic Mode : 100/0.1/0.1 (v/v/v): MeOH/HOAc/TEA
Reverse Phase : 30/70: MeOH/20mM NH4OAc, pH 5
Polar Organic Mode : 100% EtOH
Normal Phase : 30/70: EtOH/heptane
If any of them work, then I can refine the mobile phase, if not I'm onto another chiral column. There are four-five main categories of chiral stationary phases,
1. Brush type ( Pirkle ) donor acceptors - eg Whelk, (S)-tert-Leucine + (R)--(alpha-naphthyl)ethylamine.
2. Cellulose triesters or carbamates on silica - eg Chiralcel OB
3. Inclusion - eg cyclodextrins, crown ethers, polyacrylates etc.
4. Ligand exchangers - eg proline, where the mobile phase also contains a metal ion.
5. Protein - albumin, glycoprotein.
The diversity exists for a reason. I should note these columns are usually very fragile, and the wrong solvent can kill them. The best solvents for one can be fatal to another.
Derek Lowe put the following on his blog at the beginning of this month:-
" Back some years ago, the guys down the hall from me had bought one of the largest Chiracel columns that were then sold. ....This was one of the ones where the chiral packing wasn't really bonded on to anything, but just sort of layered on another powdered solid support. And as the literature included with the column made clear, this meant that you could wash the stuff right off if you weren't careful with your solvent selection.
Well, it made it clear if you, like, read the sheet and everything. Which didn't stop someone from taking up their compound in methylene chloride and pumping it right onto the barely-used $15,000 (late 1980s money) column. And in the fullness of time (say, ten or fifteen minutes), out came the solvent front from the other end: cloudy, milky, swirling with opalescent shimmers like shampoo. Which shimmery stuff was, of course, the fifteen long ones of chiral resolving agent, scoured off the packing material by the cleansing wave of chlorinated solvent.
There: clean, simple, direct, and easily avoidable by spending two minutes reading a sheet of paper. "
Your answers seem to indicate that it is always a mixed mode situation (i.e. a combination of mechanisms). But that makes for a very challenging problem for the analyst.
Can't we have - for example - a reversed phase chiral column. It seems like that would be what most folks would want.
Some of the interactions are simple, but I still can't estimate what will work with a particular enantiomer. Others might be able to, but I can't.
The problem is that most chiral columns depend on selectivity that is unique to the enantiomer structure. You can estimate whether a column will work by looking at the structure of the enantiomers in the mixture you want to separate, and often the distance of functional groups from the chiral centre can have a significant effect.
However, you have to do the chromatography to confirm. Vancomycin has 18 chiral centres surrounding three inclusion cavities - my meagre brain can't work with that.
Even the column manufacturers will ask for samples to select the most suitable column if you plan on purchasing a prep column. They can predict the likelihood of success based on the enantiomer structure, but they often want to confirm it before you make an outlay on a very expensive column, usually about 3x the price of a similar-sized high-quality reversed phase column. Such columns are not as robust as NP and RP silica-based columns.
Bruce Hamilton